totosr » Tue May 06, 2008 7:39 pm
Hello,
I have tried the cleaning procedure with 6M urea solution without results.
Moreover after washing with 75 ml urea solution followed by washing with HPLC water ca. 100 ml the pressure on the column at 0,3 ml/min was over 250 bar!
Another problem has appeared after the HPLC system was used for the cleaning procedure.
I wanted to run some samples with MS (ESI) as detector. I have changed the column to a good one. After ca. 30 min equilibration a have injected a sample and the signal was much below the yesterdays level for the same control sample. Then I have opened the source and seen that a very fine white powder in deposited in the whole ionization chamber
I have cleaned the chamber with ultrapure water but after a couple of minutes the white powder was back. It seems to block also the capillary.
I have changed the mobile phase flask back to its original after cleaning the column, and the used MF was collected in a flask during the procedure to avoid MS contamination with urea.
When using solvent from the other channel of the binary pump (but MeOH as mobile phase - tomorrow I'll try to put water on the other channel) no white powder was formed.
My question is, that is possible that urea was stacked somewhere in the HPLC system, and dilutes slowly and passes to MS. My idea is that the degasser could be the wicked peace in the whole system.
Please HELP, to get back my MS without wenting, and to get rid of the contamination.
Beste Regards,
TotosR