What do you guys think?
To not start another thread:
How important is the observation of a qualifier ion in addition to a quantifier ion in target analysis?
Best regards
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
We always record at least one qualifier ion. Recording it doesn't mean you have to check the ion ratios every time (for instance in calibration and spiked samples). It doesn't take time to record it, so better to be safe than sorry.To not start another thread:
How important is the observation of a qualifier ion in addition to a quantifier ion in target analysis?
Sometimes you just can't find an isotopically labeled version of the analyte of interest and it would be very cost prohibitive to have one custom made, therefore another functionally similar analyte that you do not expect to find in the sample matrix will work, and the closer it is in retention time to the analyte of interest the better, for evaluating matrix effects in the MS source.
Some of our European clients are now asking for quantitation to be performed and reported for both the primary and secondary mass transitions in method validations. A new SANCO guideline that I believe they are misinterpreting but what are you to do, they are the clients. For the ensuing toxicological studies that validation was done to support, that confirmation ion need not be reported other than it was there and confirmed the identity of the analyte. There have been a couple of occasions where there is no confirmation ion or it's only 10% of the primary and you can't see it in you bottom 2 standards which poses a problem. When they start talking about changing the method from a simple direct aqueous analysis to an extraction just to see that qualifier is when I just put my face in my hands.Qualifiers are not always required but the more you have, the higher confidence you can have that you do not have a false positive result.
So you have to set up your processing method to calibrate and calculate on ion 1, and again on ion 2 giving you twice the amount of data/QC/...Some of our European clients are now asking for quantitation to be performed and reported for both the primary and secondary mass transitions in method validations. A new SANCO guideline that I believe they are misinterpreting but what are you to do, they are the clients.
That is precisely what you must do. It's really no different than doing a multi-component analysis and yes you generate quite a bit of data with half of it being pointless.
So you have to set up your processing method to calibrate and calculate on ion 1, and again on ion 2 giving you twice the amount of data/QC/...
Except I'm not the one in charge of drawing lines.This is where I would personally draw the line