- Posts: 63
- Joined: Tue Oct 24, 2006 10:16 pm
Our instrument is perhaps a year old and hasn't gotten an extreme amount of use in that year. We have an Agilent GC-MS (6890/5975). Our EI source has two filaments (which can be selected from the software).
We had only used the first filament so far when last week we came across the problem where the filament just wouldn't light up. You'd hear a "click" as if voltage was been sent but it just wouldn't light up. Switching filament with the software didn't help either.
Thanks to a knowledgeable co-worker, armed with a voltmeter, we were able to test the the filaments, cables, fuses, etc. Everything seemed normal at that point, which was rather confusing. So we went deeper and started testing the power path to the filaments on the board. This is when he discovered that power appeared to go to BOTH filaments. This is to say that the relay box on the board appeared to be malfunctioning and the voltage was split and sent to both (parallel) wires at once, which in essence was not enough voltage to any of the two filaments.
The theory was verified (and the problem resolved) by removing the second filament and installing a dummy in its place.
Anyone run into this type of problem before...? Was curious as to what caused it. We typically keep the source at a temperature of 230 degC and the MS quad at a temperature of 230 degC (even in standby mode).
Thanks,
Roxanne.