I am new to this group. This is one of the best scientific discussion groups I have come across and I spent almost a day going through the past posts, and they are all really informative.
I am trying to buy a new system for our lab and was comparing the leading brands out there. I have 3 specific questions more dealing with HPLC instrumentation.
1) The representatives say that with the advanced technology now the PDA detectors have the same sensitivity as a normal uv-vis. My application, primarily with plant secondary compounds, needs a low uv detection ~220 nm. I would greatly appreciate if anyone could comment on the sensitivity issue.
2) Some of the PDA has 512 diodes and some have 1024 diodes. To me, it appears that the higher the diode elements, the greater would be the spectral resolution. But the company representatives have a different opinion. They say an element resolution of 512 diode with a 1.2 nm / element resolution and wavelength accuracy of 1 nm is good as resolution with a 1024 diode PDA. Is this right; more over does this difference in resolution make any difference in practical sense.
3) what is the main difference between a high pressure mixing HPLC system vs a low pressure mixing system. Because of the complexity of my sample matrix I usually end up with a gradient elution for analysis. Which system would be the best suited for my analysis?
I greatly appreciate any comments and suggestions.
Thank you