Seeking your valuable guidance for the development of bioanalytical method.
I am developing the bioanalytical method for glimepiride. As the drug Rt is 10.5min with developed mobile phase.
I tried the solvent for liquid-liquid extraction for extraction of drug from plasma. I tried with TBME, ethyl acetate, Combination of TBME and DCM(7:3) also with ethyl acetate and TBME(7:3).
Out of that i selected TBME:DCM(7:3). In that i observed one late eluting peak is coming to next injection and interfering with the analyte also one more co-eluting peak is also coming on the same RT of drug.
Published methods are by using diethyl ether. Is TBME and diethyl ether will make that much difference?
Also i tried with protein precipitation by perchloric acid, centrifugation followed by evaporation and reconstitution with diluent (ACN:Water 1:1). But here i am getting so much white residue after evaporation. It is not possible to inject. Also i tried protein precipitation using ethanol:ACN(1:1) followed by centrifugation and evaporation and reconstitution but iam getting one interference on analyte RT.
Glimepiride pKa 6.3 . I tried with by acidifying plasma with perchloric acid followed by liquid-liquid extraction but it is also not working the interference is there.
I optimized the rinsing solvent also so it is not carry over from syringe also.
Kindly help me out