We are using an Agilent G1888 headspace sampler and an Agilent 6890 GC. The column is an Rtx 5 from Restek (35 m, 530 μm ID, 5.00 μm film), equivalent to a USP G27. We can produce a good peak for acetone in our standards, but with acetic acid we are getting what I can only describe as "garbage" ahead of the peak and overlapping with it in such a way as to make it useless for quantitation (Figure 1 is a normal standard concentration, Figure 2 is 10X this concentration). HS sample prep is 1 mL of solution into a 20 mL headspace vial.
Figure 1: 500 μg/mL acetone and acetic acid, using headspace, in water, G27 column

Figure 2: 5000 μg/mL acetone and acetic acid, using headspace, in water, G27 column

Using the exact same GC method but directly injecting a 2 μL sample instead of using the headspace, we get the chromatogram in Figure 3, which does not exhibit the problem:
Figure 3: 500 μg/mL acetone and acetic acid, direct inject 2 μL, in water, G27 column

To see whether the use of water as a diluent could be the problem, we tried making the sample in DMF instead. As seen in Figure 4, the peak shape issue may possibly be resolved, but we see a dramatic drop in sensitivity:
Figure 4: 500 μg/mL acetone and acetic acid, using headspace, in DMF, G27 column

We tried using a different column, a ZB-624 from Phenomenex (same ID and film thickness, 30 m nominal length), equivalent to a USP G43. The resulting chromatogram of a 500 μg/mL solution in water shows some differences with the same sample run using the Rtx 5, but is still not really usable from our perspective (Figure 5).
Figure 5: 500 μg/mL acetone and acetic acid, using headspace, in water, G43 column

We also tried contacting the vendor of the headspace sampler, Agilent, for some advice. They suggested trying increasing the vial pressure to 15 psi and the vial pressurization time to 0.6 min. The chromatogram in Figure 6 resulted; this is still not an acceptable result, but there are significant changes, suggesting to us that our problem might just be a matter of finding the right headspace parameters.
Figure 6: 500 μg/mL acetone and acetic acid, using headspace, in water, G27 column, Vial Pressure = 15 psi, Vial Pressurization Time = 0.6 min

The method parameters we have been using follow. Thanks to anyone who can a) suggest improvements that might solve our problem or b) explain what is happening!!
Thanks,
Stephen
Headspace Method:
Device: Agilent G1888 Headspace Sampler
Vial Size: 20 mL
Oven Stabilization Time: 1 min
Multi HS Extr: Off
Extractions Per Vial: 2
GC Cycle Time: 26 min
Inject Time: 1.00 min
Loop Equilibration Time: 0.10 min
Loop Fill Time: 0.20 min
Loop Temperature: 95 °C
Oven Temperature: 80 °C
Shake: LOW
Transfer Line Temperature: 105 °C
Vial Equilibration Time: 40.0 min
Vial Pressurization Time: 0.40 min
Carrier Pressure: 4 psi
Vial Pressure: 10 psi
Agilent 6890 GC Method:
Oven Program:
Initial Temp 40 °C, hold 0 min. Ramp at 15 °C/min to 100 C. Ramp at 20 °C/min to 200 °C. Run time 13 min.
Inlet:
Mode: Split
Temp: 140 °C
Pressure: 4.26 psi
Split Ratio: 9.99:1
Split Flow: 48.4 mL/min
Total Flow: 55.7 mL/min
Gas Saver : On
Saver Flow: 20.0 mL/min
Saver Time: 2.00 min
Carrier Gas: Helium
Column:
Mode: Constant pressure
Pressure: 4.26 psi
Nominal Initial Flow: 4.9 mL/min
Average Velocity: 35 cm/sec
Detector:
Temperature: 250 °C
Hydrogen Flow: 30.0 mL/min
Air Flow: 400.0 mL/min
Mode: Constant makeup flow
Makeup Flow: 25.0 mL/min
Makeup Gas: Nitrogen