by
1D1P1J » Thu Jan 25, 2018 10:31 pm
Thank you for all of the replies.
Firstly - I have solved my original problem - the small pieces of capillary tubing were attached to lengths of column that had broken. I've removed the ferrules.
Here is a picture of the Dean's switch. It is from Thermo Fisher.
I'm confident with DIY gas plumbing.
It is my understanding that a Dean's Switch with two auxiliary gas flows can be used to split the from from a single column onto two detectors. The extra flow to the MS vacuum is counteracted by attaching a length of narrow bore capillary column from the Dean's switch to the MS. The fine control of the flow split is then provided by the auxiliary gas flows.
The application is straightforward: FID quantitation with simultaneous MS peak identification. This allows us to adhere to certain industrial standard methods, while also generating MS data to ID unusual peaks that pop up from time to time. It also allows us to use direct MS quant for other applications.
@ GOM: That low dead volume, two hole ferrule sounds like a much simpler solution, but I think the Dean's Switch allows more flexibility i.e. if the MS vacuum is higher or lower on a given day, the gas split can be controlled by adjusting the aux gas flows to the Dean's Switch.