What are the symptoms that a trap is dying?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Purge-trap system w/ Aglent GC-MSD

all of the sudden my early 'gas' peaks are broadening out. I checked to see that all fittings are tightened but I still have this problem.

I've had the trap for a year and I've run lots and lots of samples.

I'm using a vocarb k trap on a tekmar concentrator...

oh, and one other question...

do you use new ferrules when changing the trap? I've bought the new trap and didn't know that I needed furrules. If I'm careful can I just swap the old trap for the new? or should i order new ferrules and be down for a couple of days?

Thanks.

We use an OI 4560 concentrator with a number 10 trap. Typically when the trap is starting to fail the ketones, carbon disulfide and especially bromoform will respond poorly.

Regardless we change our traps every month.

broadening peaks... do you think you might have a slight leak? check the concentrator (we once had a similiar problem and found a slight leak coming from the interface)
3 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Amazon [Bot], Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry