Hi everyone,

I am pretty new in gas chromatography so i would like to ask you few questions ;

I have HP 6890 Series GC system with MSD Detector 5973 Agilent and I have installed capillary column SupelcoWax- 10 with dimensions 30 m x 25 mm ID x 0,5 um, cause i would like to separate monoterpenes and higher alcohols and i have found column bleed.

Is this right choice of column for GC-MS?

Choice of column is dictated first by what you need to separate, and then by which detector you have. How much bleed you can tolerate depends of the size of your peaks. In general wax phases bleed more than non-polar silicones so if you have to have minimum bleed and you can do your separation on a non-polar column switch to a silicone.

Wax phases are very sensitive to water and oxygen in the carrier gas, install a fresh gas scrubber and check carfully that you have no leaks in the plumbing and the inlet.

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry