jdlh199 » Tue Mar 18, 2008 6:56 pm
I am still of the opinion that the 'control of bias' applies to the study results and not to the validation of the analytical method.
Yes that's right, this is for a GLP sample analysis study using a method that has already been validated.
With a valid analytical method there are checks and balances at every turn which do nothing else but guarantee the validity and lack of bias in your data.
Chances are with ever analytical sequence you produce you will have:
Proof of system functionality from the system suitability test;
Linearity, precision and accuract within criteria from the calibration data.
Intra run accuracy (and possibly precision) from QC analyses.
all of which must meet specified criteria for the sample results to be considered valid. This is your demonstrable control of bias
I think that's kind of what I thought as well, perhaps our QA is being insistent in simply requiring a statement that says "this will reflect control of bias" with those magic words in the protocol. At the moment it just says something along the lines of "calibration data must be within XX, QC data must be within XX etc etc" without mentioning "control of bias"
I think all I need to add is a simple statement along the lines of "Control of bias is assessed through the accuracy (%recovery) and precision (%RSD) of QC data at each level". For the study report following analysis I can then just show the mean and %RSD of each QC level as an examination of control of bias.
That'll do.....
Thanks everyone for the comments!