- Posts: 15
- Joined: Sat Jan 05, 2008 11:04 am
I am running a headspace residual solvents method on an Agilent 7694A (G1289A) headspace autosampler attached to a 6890A GC. I have connected the column directly to the transfer line ZDV union which gives good peak shape in the standard. However the RSD of the analyte peak areas is around 15% and the RSD of the internal standard ratio is >50%. The internal standard is acetone and I am quantifying chloroform. The sample matrix is purified water for the standards.
If I connect the headspace to the split/splitless gas line and run the inlet in splitless mode (no purge and capped septum purge) with a 1mm ID liner I get very good (approx 3% RSD) peak areas and <1% RSD Int Std ratios. But with this setup I cannot meet some of the method suitability parameters (system sensitivity SNR).
GC Parameters
Column: RTX-WAX (60m x 0.32mm, 0.5um film)
Detector: FID @ 280DegC
Inlet: 110DegC
Oven Temp: 40DegC initial
Does anyone have any idea what is causing the large injection to injection variability in the peak area when the headspace is hooked up directly to the column? Any help would be apprecated as I have been working on this for weeks!