What mechanism can cause this?

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I made a 50% toluene 50% tetradecane by weight standard, mixed, and got 27% tol and 73% tetradecane by Area% on my 5890 Series II TCD detector!!! I shot the same mixture on a 6890 with same conditions and got 52% toluene and 48% tetradecane by Area% on its TCD (which is what I expected). Has anyone ever seen this type of discrepancy or have any idea what is wrong with the 5890?

Thanx for your time.

My money would be on inlet discrimination.

Peter
Peter Apps