GCMS troubleshooting...

Posted: Mon Feb 18, 2008 9:20 pm
by Goonie
For about a week now I've been experiencing what seems like a leak, although not sure.

My guess is that I have a minute leak in the transfer line into the MS.

But my chromatorgraph looks like I might have a bad contimination. I changed the helium tank when this started to occur, and unfortunately I changed the column at the same time, so I cannot narrow it down to just the tank. I'm also thinking I might have introduced oil from the tank into the system. I've cleaned the source to see and only a slight improvement.

Just not sure, thought I'd ask here first. I call this the "grass effect" as though the peaks come out fine but you have these plateous randomly.

Image
Image

Posted: Thu Feb 21, 2008 3:22 pm
by Goonie
Ok, so I believe I have figured it out. I switched the rough pump and am getting a great baseline. Must of been leaking oil into my MS. So far so good.


Goonie

Posted: Fri Feb 22, 2008 7:23 am
by Peter Apps
Hi Goonie

Thanks for updating us, are you running full scan, or SIM ? If SIM, do the steps in the baseline coincide with changes in the ion(s) that youare looking at ?

Peter