For about a week now I've been experiencing what seems like a leak, although not sure.

My guess is that I have a minute leak in the transfer line into the MS.

But my chromatorgraph looks like I might have a bad contimination. I changed the helium tank when this started to occur, and unfortunately I changed the column at the same time, so I cannot narrow it down to just the tank. I'm also thinking I might have introduced oil from the tank into the system. I've cleaned the source to see and only a slight improvement.

Just not sure, thought I'd ask here first. I call this the "grass effect" as though the peaks come out fine but you have these plateous randomly.

Image
Image

Ok, so I believe I have figured it out. I switched the rough pump and am getting a great baseline. Must of been leaking oil into my MS. So far so good.


Goonie

Hi Goonie

Thanks for updating us, are you running full scan, or SIM ? If SIM, do the steps in the baseline coincide with changes in the ion(s) that youare looking at ?

Peter
Peter Apps
3 posts Page 1 of 1

Who is online

In total there are 197 users online :: 1 registered, 0 hidden and 196 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: Amazon [Bot] and 196 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry