Large Peaks
Posted: Sun Sep 26, 2004 2:02 pm
by Fabiano
Some people say that the most retained the peak more large it is.
I say that I believe that it is more large due to the low efficience of %B in my partition process. but they state that is because of Diffusion, only this.
who is right?
Posted: Mon Sep 27, 2004 3:14 pm
by tom jupille
Diffusion essentially controls how much your sample band or zone broadens as a function of migration distance. To a first approximation, every compound in your sample has the same zone width when it reaches the end of the column. The observed width of a peak depends on the width of the zone when it reaches the end of the column, and also on the distribution coefficient (more strongly-retained peaks are eluted from the column in a larger volume of mobile phase).