Large Peaks

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Some people say that the most retained the peak more large it is.

I say that I believe that it is more large due to the low efficience of %B in my partition process. but they state that is because of Diffusion, only this.

who is right?

Diffusion essentially controls how much your sample band or zone broadens as a function of migration distance. To a first approximation, every compound in your sample has the same zone width when it reaches the end of the column. The observed width of a peak depends on the width of the zone when it reaches the end of the column, and also on the distribution coefficient (more strongly-retained peaks are eluted from the column in a larger volume of mobile phase).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 306 users online :: 2 registered, 0 hidden and 304 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 304 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry