These baselines were produced using an Agilent 1100 system, a DIONEX CS12 4x250mm column and guard column, a DIONEX cation suppressor, and an old Waters 430 conductivity detector. The solid baseline is water, the large dashes being 0.04N H2SO4, and the small dashes with the peak is a 2.3 ppm ammonia standard. The baseline dip becomes much more pronounced with the presence of ammonia.
I'm thinking that the injector is somehow involved since the baseline dips are occurring in the same places each time. But the Agilent injector is a flow-through design, so if it's contamination, the valve would be the first place to look, right?
This doesn't happen when I use our DIONEX system to perform the analysis. I think we may need both systems to get through a large set of samples, so any help is appreciated.