Imp quantification

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi all,

Quick question if I may:

What are your views and those of the FDA etc on quantification of imps by RRF?

Specifically - would you quantify imps based on external standard (of the main compound) with an RRF. Or would you just quantify off the main peak in your actual sample and again apply the appropriate RRF?

I would assume the former is the 'most' GMP but is the later just as valid?

Ultimately...for GMP analysis do you run standards or just SST, Blank, LOQ and samples?

Many thanks,

Serp

We just quantitate against the main peak by % a/a and apply RRF of impurity peaks if we happen to have them.

As far as I am aware this is pretty much standard practice in pharma.

Using this approach you would only need to run external standards if you were also looking to aqcquire assay content data. Prefrably I run a separate fast LC method to acquire content assay data.
2 posts Page 1 of 1

Who is online

In total there are 381 users online :: 3 registered, 0 hidden and 378 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google [Bot], Majestic-12 [Bot] and 378 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry