I am new to forum and new user for HPLC. Can some one help me in establishing protocol for validation of invitro samples and biological samples. I will give my protocol and please correct me where ever it is necessary at the same time update me with new regulatory requirements and limits.
1) Accuracy: Lower, intermediate and higher concentrations of analyte were prepared (n = 5). Measured as % Relative error.
2) Precision: Lower, intermediate and higher concentrations of analyte were prepared (n = 5). Measured as % Relative standand deviation. For intra-day and inter day varibility study n = 6 samples per day for 3 days.
3) Specificity and selectivity: Inject blank (n = 5) for checking of interference at analyte retention time.
4) Linearity range: Five different concentrations (n = 6). Apply one-way ANOVA for linearity checking.
5) Ruggedness: Change the instrument, analyst, column, brand of solvents and inject Lower, intermediate and higher concentrations of analyte (n = 5). Report as % Analytical recovery ± SD.
6) Robustness: Change the pH of mobile phase (± 0.1) and inject Lower, intermediate and higher concentrations of analyte (n = 5). Report as % Analytical recovery ± SD.
7) LLOQ LLOD: ?????????
8)System suitability: How to do it?????????????
Sorry for such long question. Expert comments will help me in designing protocol.