Hi everybody! :D

I have a big, big problem! Can you help me to find some stability indicating method hplc for quantitative analysis of allantoins impurities in losions and creams. I tried using amino column and hilic silica column, but I did not get satisfactorily results, because I didn't see impurities and peaks of allantoin and urea get out together.

Thank you!
Best regards
Mila
mila

Hi Mila -

This looks to be a difficult assay. Allantoin and urea
on Unison UK-Silica:

http://www.silvertonesciences.com/files/TI249E.pdf
http://www.silvertonesciences.com/files/TI178E.pdf

A quick search on google came up with an article where
the authors used an NH2 phase and 70:30 acetonitrile:water.

Did you try an NH2 phase? What did your chromatogram look like?

Dear Mila,

Today you may find several HILIC columns commercially available, most being either plain silica, or amino phases. These columns have limitations as the strong ionic interactions will contribute substantially to the separation mode. This is well described in recent studies, and below is two good references in comparing HILIC columns.

Retention Behavior of Small Polar Compounds on Polar Stationary Phases in Hydrophilic Interaction Chromatography
Y. Guo, S. Gaiki
J. Chromatogr. A, 1074 (2005) 71-80

Investigating the Effect of Chromatographic Conditions on Retention of Organic Acids in Hydrophilic Interaction Chromatography Using a Design of Experiment
Y. Guo, S. Srinivasan, S. Gaiki
Chromatographia, 66 (2007) 223-229

To obtain a different selectivity, you should aim at using a bonded phase, and I would suggest our zwitterionic phases as it gives a much different selectivity than all other HILIC phases.

Being a permanent zwitterion, it is charged yet neutral overall, which allows the use of much lower ionic strength in the mobile phase due to weak electrostatic interactions which is contrast to the strong interactions obtained with the two other alternatives.

Being a permanent bonded zwitterion, it is pH independent allowing the use of experimental conditions set for the analyte rather than for the column.

I would suggest you check in at our website for further reading or contact me via www.sequant.com
Merck SeQuant AB
www.sequant.com

Looks like this has to be repeated: If you can´t use the ideal pH and/or ion concentration in your mobile phase, then the ZIC-HILIC does indeed show strong pH dependence.
4 posts Page 1 of 1

Who is online

In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Adsense [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry