A know two types of Agilent autosamplers The G1329A (or B) is the older ALS 100 vials, good RSD and carry over. You can wash the needle from a certain vial position. No other outside wash method. You can set injector programs to do some mixing, dilutions, valve cleaning, etc. Easy to extend the standard 100uL max inj. volume to 900. Termostatted from 4 to 40 celsius. We had 1100 models, but the same type exist in 1200 brand.
The other G1367A (orB) is the well plate sampler can handle plates and vials as well. In some of my applications has better performance then the 1329, in others appl. the same. This can handle 100, or 118 vials, and a lots of plates (Ihave no experience). WPS has the vial-used needle wash (as the 1329) and has a "Flushport" function, which uses a peri pump for the needle wash in a port. This has an other usefull function, the "well bottom sensing" for less amount of samples. This type can reduce the delay volume for fast gradient applications. This type has a 4 steps injector valve cleaning setting , but I never used. Basicly up to 100 ul, but theoretically extendable up to900. This has a "micro" version, but that is only for capillary flow applications. Termostattable, 4 to 40 cels.
As far as I know the A types up to 400 bar, the pressure limit forB types is 600 bar.
A nice difference at the first sight: The 1329 carries the vial to the needle, and in the 1367 the needle goes to the vials.
This were the analitical samplers
Could I help? Attila