ece » Wed Dec 19, 2007 3:35 am
the astm method specifies 3 ml/min and it doesn't specify the column length/diameter. why is 3 ml/min too low a flow rate for that diameter?
the peaks i have been getting (including the heptane peak) have been very sharp and narrow, which i attributed to the flowrate being too high. when i lowered the pressure even more, the heptane peak achieved some of the broadness that it had before.
why would you inject lighter gas? you mean like methane or propane? if it's a gas, how is it going to retain onto anything? it seems like it would simply flow through the column, as the carrier gas would. so...why not simply measure the carrier gas flow rate directly and do a pressure drop calculation using the length/diameter of the column, then determine if it fits that linear velocity that way? if retention time is necessary factor, why not just insert a blank solvent and determine it's retention time along with another blank solvent and determine it's retention time and average them? Then you can determine the mean linear gas flow rate for the flow rates for the substances you are determining?
i suppose it is a shorter column with a larger diameter than we were using before, but the method, although it is poorly written, says use 3 ml/min as the carrier gas flowrate and doesn't specify the column length/diameter as a corresponding value.
it makes sense that at that pressure the EPC is not able to control the pressure adequately at that low of a pressure, i've seen that i've had to make a lot of adjustments and it does not adjust for approximate values when it is that finite.
it also makes sense that the mean linear velocity of the gas through the column is incorrect, and that the method was simply poorly written and assumed all columns were the same length/diameter and so did not specify that. even the standard deviation calculation for the calibration has a typo in it that renders it incorrect, so i believe it is possible that they simply did not take carrier gas flow rate as it relates to column diameter, pressure drop, and column length into account.
if the flow rate through the column is too high, what typically happens? do the peaks get narrower as i predicted, or is it something else? can the detector still pick up the peaks at the appropriate concentrations when the flow rate is too high?
-thanks all for your input. sorry for my doubts, i tend to question everything. i am a beginner with gc. thanks