Can I stwich between buffers in 5min in my program?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

14 posts Page 1 of 1
I have the following gradient program:

Time Water% ACN% 20mM kH2PO4 (pH3)
0 50 50 0
11 0 100 0
25 0 100 0
30 50 50 0
30.01 0 20 80
35 0 20 80

Stop 35min
post time 5min

initial condition
Time Water% ACN% 20mM kH2PO4 (pH3)
0 50 50 0

at 30.01 min, I switch from 50:50 water/ACN system to 20: 80 Buffer/ACN system and at 35 min, I change it back to 50:50 water/ACN system.

That is unusual gradient program for me. I am curious, why do you need such switching? Is there any particular reason for this?

I did compose all kind of gradients so far and I do not think it is unusual. The only concern I do have about this is the combination AcN-Phosphate, can you eliminate the possibility that there will be no precipitation of phosphate? I mean, the phosphate concentration is low but I saw some precipitation on that level to. However, the gradient can help providing some specificity in separation.
regards
goxy

I did compose all kind of gradients so far and I do not think it is unusual.
Ok, then please give me reasonable explanation to me what is the role of 5 min change of buffer instead of water (when the mobile phase of the whole run is consisted of ACN and water)?

Well, I suggest to perform some experiments on column re-equilibration effects when using water and phosphate buffer. My observation is that the column re-equilibration is significantly faster when using inorganic part of high ionic strength instead of (acidified) water. I do not know why there is a change from water to AcN here, but the re-equilibartion can be a reason.
regards
goxy

My observation is that the column re-equilibration is significantly faster when using inorganic part of high ionic strength instead of (acidified) water.
Ok, I haven't done such experiments so I am not sure whether this is true or not (I'm kind a suspicious about that). Maybe other members can give their opinions regarding this issue. All I know, regarding the time needed for column re-equilibration, is that the greater difference between the mobile phase composition at the end of the run and the initial mobile phase (at the beginning of the run) the longer the re-equilibration time. The duration of the column re-equilibration process is affected also by the value of the gradient dwell volume and column void volume.

Regards

I do agree that the column equilibration is affected by the facts you wrote. However, I will take care to post some publications describing these effects. However, give me some time since I am quite busy now. Many tips about this topic can be found in LCGC.

Regards
goxy

Thanks for your replies.
The reason is because I would like to flush a retained component out at the end of the run (after 30min). I do not care about the sepeartion of that component. I only care about stuff coming out before 25 min. And I do not want the mobile phase changes to affect my separation before 25min, which means I hope the re-equilibration can be done in 5min (post run 5min).

I understand another choice would be to use a guard column.

By the way, no instant precipitation was observed, but I am not sure if there is slow effect.

Please give me some suggestions about the following issue.

About my retained components, I have 3 options:
1) flush it out for each single injection (I wasted 5min to flush out retained components, and I do not know if 5min is enough to equilibrate between mobile phase change)
2) flush it out for every N injections (I can save some time, and re-equilibration time can be set longer than 5min. But is this practical? I might have ghost peaks)
3) use a guard column (question: what about those components I am looking at? Retention times of them will change, and maybe I cannot save much time.)

Moonchips,

You should post the column dimensions and flow rate in order to expect any suggestion regarding equilibration time.
One thing you can be sure of: A guard column will not speed up the equilibration. On the contrary!

Best Regards
Learn Innovate and Share

Dancho Dikov

Rule of thumb for re-equilibration is 3X system volume + 5 X column volume. Unless you have a very short column or very high flow rate it is probably longer than 5 minutes.

Dwell volume is 0.8mL. Column is 150mm*4.6mm*5um. I am not sure about Column volumn (maybe about 70%?)

What is system volume?

Moonchips,

The described column has approximate volume of 1.75 mL. (refer to equation 1)
If the flow rate is 1 mL/min, it’ll need 11 – 12 min equilibration time.

(1) Volume (column) = D^2 x 3.14 x L x f/4

D = column diameter in cm
L = column length in cm
f = approx. 0.7


Best Regards
Learn Innovate and Share

Dancho Dikov
14 posts Page 1 of 1

Who is online

In total there are 17 users online :: 3 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahmadseo, Ahrefs [Bot], Bing [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry