Rajdeeproy,
I would not necessarily say that AZ would limit sensitivity. However, if I assume that your full signal out is 1 Volt then you have eaten up at least 10% of you linear range if not more. Furthermore, if you normally run at a few mV as I suspect, the real issue is that your detector is probably dirty and THAT would cause a loss of sensitivity. You might try hydrogen cleaning your ECD if there are instructions on how to do that or you might send you ECD off to a reputable shop for cleaning or to PE for cleaning. I am not familiar with PE ECD but when the HP ECD is running about 10% of full scale, it desperately needs cleaning. Sorry bud.
Best regards.