- Posts: 7
- Joined: Wed Nov 14, 2007 7:00 pm
I new to LC. I run my samples which should contained enough target compound,with the new replaced column. My samples did not run before. And conditions and phase were followed by the papers.
But no peaks coming out. Baseline fluctuating +-5. I also did not see the injection peak.Shoud that always appear?
Many problems may have:
1) column leak?
2) needle problems?
3) my lamp dead?
Could you guyes tell me the most likely culprit? In which priority,I should trouble shooting this problem. Any procedule I can use to diagnosis this possibility?
========
Agilent 1100 HPLC system
Column: Phenomenex (Torrance, CA) Curasil PFP 15 cm, 2 mm, 3 µm column
Injection volume: 10uL
Mobile phase: water-acetonitrile started 35% ACN increasing this linear to 60% in 6 min. Both solvents contained 200 µL/L trifluoroacetic acid
The wavelength 304 ± 4 nm
==============
thank you
creageng