XL » Sun Nov 18, 2007 7:43 am
I worked with organic acids for some time in the past, and have collected some information on this subject.
In general, the following methods can be considered for this purpose:
1. Anion-exchange IC method using suppressed conductivity detection and IonPac AS-11 HC
http://www1.dionex.com/en-us/columns_ac ... s3609.html
2. Ion-Exclusion LC using UV detection or IC method using suppressed conductivity detection and IonPac ICE-AS1 or AS6 columns
http://www1.dionex.com/en-us/columns_ac ... s3630.html
3. RPLC method using UV detection in ion-suppression mode
http://www1.dionex.com/en-us/columns_ac ... s6953.html
4. Mixed-Mode Chromatography LC method using UV detection (see Figure 7 in
http://www1.dionex.com/en-us/webdocs/48 ... 021407.pdf)
Usually the K' of acetic acid on a RP column is around 1 to 1.5 as ljc suggested above. Considering the application requirement and complexity of the sample , it is often not adequate. For this application, I believe Acclaim Mixed-Mode WAX-1 column is a good candidate (see the last link above). Chromatographic conditions include: Column dimension, 4.6x150 mm; mobile phase, 25 mM phosphate buffer, pH6; flow rate, 0.8 mL/min; UV at 210 nm. The K' is around 3. The interference from sample matrix can also be minimized on this column.
You are welcome to contact me regarding this application.