2 ml/min for a 3um particle size column?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Greetings.

We are validating a method transferred from another company, which has been approved by FDA
they use ODS inertsil 10cm*4.6mm*3um column. the column flow is 2ml/min. is it too fast?
how is column flow related with the particle size d ID?

Thank you in advance!

It's OK if your pressure stays OK and doesn't overpressure.

It shall be OK. The column is kind of short.

how is column flow related with the particle size d ID?
Flow rate is related with the column internal diameter:

Flow rate = f (diameter^2) ; f - function

e.g. if the column I.D is 4.6mm at 2mL/min flow, then the flow that should be used if you change the column with another with 2.1mm I.D is:

(4.6mm/2.1mm)^2 = 4.8 (rounded to 5)
The flow rate will be 2mL/min:5 = 0.4mL/min for the narrower column.

Regards

I have run many methods on a 4.6 x 10 x 3.5 micron column at this flow rate. For water, I expect a pressure around 3000 psi, less for mixtures of acetonitrile and water. Only for a methanol-water mixture, you are pushing the limits.
5 posts Page 1 of 1

Who is online

In total there are 33 users online :: 1 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 32 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry