Diagnosis for system block procedure????

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
I am a new HPLC user. My Agilent 1100 system was blocked. I listed how to diagnose system block as below. Please add more that I missed and correct any mistake I made.

How to solve frit block problem?
How to do pressure test?

Diagnosis for system block procedure:
1. switch a column to a connecter to check if the pressure can drop. Yes, Column blocked. No, it might not be the column
2. open the connecter to see if there is any liquid coming out (1ml/min). Yes, it is after the column. No, it is before the column
3. Open the black valve. With 1ml/min flow, if the pressure is larger than 5 bar, the frit is blocked (the salt concentration of my mobile phase matters)
4. 6-port valve can be used to diagnose the block location

You are on the right track. The general rule is to start disconnecting fittings from the end (downstream) of the system and then work your way upstream. When you disconnect a fitting and the pressure drops, your blockage is just downstream from that fitting.

You can sometimes backflush (reverse flow) a plugged frit, but there no guarantee that it will work. It is usually more cost-effective to simply replace the frit.

Details on pressure test procedures vary from instrument to instrument. Your Agilent 1100 manual should include the pressure test procedure.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Thanks a lot, Tom

Ziminali-Did you solve your problem?

As a preventative you can typically flush with water if you are using water soluble solvents (for example acetonitrile or methanol) on the system. Combinations of high concentration salts (for example phosphate buffers) and organic solvents will precipitate out in the system. A good rule of thumb is to replace the buffer with water and flush out the column/HPLC. See p20 to replace the frit.

http://www.chem.agilent.com/cpdocs/1100 ... enance.pdf

thank
I found the location.

it is the 3 ul heating part that is blocked. I was running 50 C and I stopped flow without lowering the temperature first. so that part ran dry and some salt condensed and blocked the path right away.

thanks everyone again.
5 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry