this may be a silly question but I have spent a couple of hours searching the forums and haven't got the answer I need yet. I have been assaying for catecholamines for the past few years and have recently tried a new method to increase my selectivity and sensitivity but have a question regarding its improvement. Currently, I run an agilent 1100 system with a coulochem III ECD and a sunfire C18 RP column 4.6x100; 5um packing @ 1.2mL/min flow rate. My mobile phase is as follows:
75mM NaH2PO4
500uM EDTA
5mM Octane sulphonic acid
18% ACN
pH 5.25
The issue is there are unknown peaks that elute around my peak of interest in a real sample. I originally ran lower ACN and lower Octane concentration but have found that increasing both keeps my amines at a similar retention whilst decreasing the retention of the unknowns (hoping to wash them out before my peaks of interest elute). I was pondering whether to replace some ACN (say 5%) with MeOH given that this should bolster the effect of the ion pairing reagent, is this a correct assumption? see: http://www.sepsci.com/chromforum/viewto ... etonitrile
So my questions is, are there any problems using a combination of ACN and MeOH rather than just switching from one to the other?
Thanks for your help and sorry for the long post.
Cheers,
James