Dear Members,

I would like to hear from you a suitable solvent pair ratio for the analysis of Flavonoid glycosides from leaf and seed samples. I am using Shimadzu LC10A with PDA. I have tried with ACN:water 70:30, but no proper elution of peaks I could notice.

Please inform me any suitable method avialable for the analysis of crude leaf extract for flavonoid cojugates.

Best regards

Best approach would be to start with a gradient covering a wide range of ACN concentrations (e.g. 5 - 95%).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Here is an example of some examples of flavonoids
(Imtakt used a gradient with TFA as the acidic modifier):

http://www.imtakt.com/TecInfo/TI294E.pdf

You will want a gradient method. I did anthocyanin conjugates in bilberry with a 5-95% MeCN gradient and had peaks almost the whole way across. http://www1.dionex.com/en-us/flash/fullflash62071.html I also tried MeOH and got somewhat different selectivity. The choice of acid is not that critical as long as you can get down below pH 3. A PDA is really nice to have too.
Mark Tracy
Senior Chemist
Dionex Corp.

Echo Mark's comments on pH and PDA, however the choice of acids might have a small effect on your sensitivity.

It won't bother your flavonoids so much, but other components that don't have strong chromophores and which might co-elute with your flavonoids will be easier to detect if using a PDA.

Phosphate absorbs less than acetate or formate in low UV so your gradient "ramp" will be less significant. If using an acetonitrile based MP, I might be tempted to use dilute phosphoric acid first to keep the baseline as flat as possible, especially across a broad ACN concentration.

Best of luck!

CJ

Also, a cool site for you to look at re: flavonoids:

http://www.metabolome.jp/software/FlavonoidViewer
We have used the following method in the analysis of flavonol glycosides.

Column: Phenomenex Prodigy ODS 5u (3) (250 x 4.6, 5 um) with pre-column Phenomenex Prodigy ODS 5u (3) (30 x 4.6, 5 um)

Detection at 270 nm and 370 nm with a PDA detector

Solvent A: water-THF-TFA (98:2:0.1)
Solvent B: ACN

Solvent gradient:

Time (min): proportion of solvent B (%)
0-2: 15
2-14: 15-25
14-19: 25
19-24: 25-60
24-28: 60
28-30: 60-90
30-35: 90
35-40: 90-15
30-40: 15

In my colleague´s experience, flavonol glycosides should not be dissolved in ACN before injection, since they are not stable in the solvent. you might want to try e.g. MeOH-water (1:1)

Here's another application

Quercetin glycosides separated on Unison UK-Amino:
http://www.imtakt.com/TecInfo/TI369E.pdf
8 posts Page 1 of 1

Who is online

In total there are 11 users online :: 2 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry