Removal of SDS in order to analyze by LC/MS
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
I am currently analyzing an ophthalmic drug product that has been photodegraded in order to generate an impurities profile. I'd like to collect fractions of several of the observed peaks for LC/MS analysis. However, my diluent and mobile phase is a 58% aqueous solution of sodium dodecyl sulfate with 0.6% H3PO4 mixed with 42% acetonitrile. Is there any extraction technique available to removed the SDS in order to make the samples more LC/MS friendly?
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
If you plan to re-analyze by LC/MS then you can just let the LC separate them. If you plan to just do an infusion into the ESI source, then you need to remove both the SDS and phosphate. For that, try a weak anion exchange SPE cartridge. Use a compatible volatile acid like formic for elution.
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
2 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 16 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 16 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science