What is the difference between EPA 524.2 and 8260?
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
bdh,
Back when I routinely did this sort of thing....
Among others, 25 mL sample volume instead of typically 5 mL sample volume and use of a frit sparger instead of a needle sparger. Which means lots and lots of water on your trap. Use of a single internal standard instead of three, three point calibration instead of five, much lower detection limits (more about the 25 mL than at the instrument), and others.
Biggest issue I came across was the impact of so much water on the trap and analytical system, even with split injection.
Have you been able to get your hands on the method?
Best regards.
Back when I routinely did this sort of thing....
Among others, 25 mL sample volume instead of typically 5 mL sample volume and use of a frit sparger instead of a needle sparger. Which means lots and lots of water on your trap. Use of a single internal standard instead of three, three point calibration instead of five, much lower detection limits (more about the 25 mL than at the instrument), and others.
Biggest issue I came across was the impact of so much water on the trap and analytical system, even with split injection.
Have you been able to get your hands on the method?
Best regards.
-
- Posts: 8
- Joined: Thu Apr 12, 2007 9:08 pm
I am currently utilizing Method 8260....I just recently have had a few people inquire about EPA 524, so I thought I'd look into it. I can't seem to find a copy of EPA 524 though.
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
BDH,
Try:
http://www.accustandard.com/asi/epa_downloads.php3
As usual, the devil is in the details.
Best regards.
Try:
http://www.accustandard.com/asi/epa_downloads.php3
As usual, the devil is in the details.
Best regards.
-
- Posts: 8
- Joined: Thu Apr 12, 2007 9:08 pm
Thanks!!
5 posts
Page 1 of 1
Who is online
In total there are 129 users online :: 0 registered, 0 hidden and 129 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 129 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 129 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science