Irregular peaks

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
HI!

I am new to this forum, but I take the chance.

I am developing a new method for an endogenous compound. My problem is that the peaks are quite irregular. The AUC seems quite consistent when running the same sample over and over, but i am still concerned that this would lower the sensitivity and precision. Both low and high concentrations peaks are irregular, and the "right" side is the worst. Some minor tailing iss also present. Assymetry factor about 2.

I am running a gradient program in ACN 15-70% (with 0,1% formic acid) over 6 minutes. The column is a C18 ID 2.1 50mm. Detector is a mass spectrometer.

Any clues? Suggestions?

Thanks!

First of all, an asymmetry factor of 2.0 is not "minor", especially in a gradient.

Secondly, with a mass spec, it may not matter, especially if you are getting reasonable quantitation.

It is difficult to say anything more specific without more details an, ideally, a chromatogram to look at.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry