rubyfan » Tue Sep 18, 2007 9:43 pm
Has anyone here done any negative inlet pressure GC-MS? I work for an R&D company and we're trying to do some testing of a 1.5 meter, 0.25mm ID analytical column. I have the column connected to a 1079 PTV and plumbed directly into the Saturn 2000 mass spectrometer. For my initial start temperature of 85C, the inlet pressure is at -9.3psi. I found initially that when the split comes on at 0.75 minutes into the run, at 100:1, the inlet pressure would jump way up to over 1psi. So I changed the method to keep the split closed through the analysis. That still hasn't worked. Now with the oven at 190C, the inlet should be at -7.9psi, but it's reading 0psi. So, I'm having difficulties maintaining the negative pressure. Seems fine for the first several minutes, then it can't maintain.
Also, I notice a lot of air/water with the negative pressure.
I just need to do a couple of quick tests with the instrument configured like this. Any comments or suggestions? I could try installing the column into the front injector, which is an 1177.
-Aaron