Weird Baseline

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
we had a baseline problem this morning in agilent 1200 series with DAD. the baseline was recorded as in the picture:

Image

the 'wave' was not interfered by flow. though the flow was stopped, the 'wave' was still appeared. it could be eliminate after we shut down the system and then turn it on. I conclude that the source of the problem is in detector but I have no idea what it is exactly. could you please give me any reason about this phenomena as the basic to build a reasonable preventive action?
thanks in advance.

regards,
siswanto

Two thoughts...

1) How old and how many hours are on your lamp?

2) When the detector is on, is the fan running consistently, or does it cycle on/off?
Thanks,
DR
Image

another thought or comment

some years ago I've experienced a comparable phenomenum:

it was caused by the airco (power supply) . the pattern in the detector signal equals the on-off of the airco. .....

but here the time between the peaks is rather short , but who knows....
3 posts Page 1 of 1

Who is online

In total there are 11 users online :: 2 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry