Can anyone tell me the proper settings for a a HP5890 GC Series II GC split injection port with respect to: injection port temperature, split valve on/off time, scan rate, and proper flow. We run volatile methods by EPA8260C so we'd be looking for early eluting gases (bromomethane, chloroethane, etc) and late eluting dichlorobenzenes and naphthalene. Right now, we have a Agilent DB624 column 25m .2mm ID and we have the injection port temperature at 210 deg, the split ratio at 30:1 (flow into the column at 1ml/min) and the split valve scheduled to turn on at 1 minute and off at 2 minutes (we are interfaced with a Tekmar 3100 Purge and Trap Unit to transfer sample to the GC). Basically, we are getting peaks that either show signs of fronting or tailing. We've tried to adjust flow with no great success either way.
Also, with regard to tuning, after source cleaning, we get good ratios 69/219 at 50% and 69/502 at 3% with EMV at 1700. After a few days, tuning becomes erratic (can't get good ratios) and EMV jumps to 1900+. It would indicate a dirty source, and indeed after a few days, I took the source out and there was an inordinate amount of burn on it, even more than you would expect from fairly aggressive tuning over a 3 day period. Anybody have any ideas as to what the issue would be with that? I should also mention that pervious to this my boss had suggessted leaving the split vent valve constantly "on", so I don't know if that would be contributing to the dirty source. I know this is a lot, so any help would be greatly appreciated.