Post-column derivatization system hight pressure ploblem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi to all!
We have a HPLC system with a post-column derivatization system from Pickering laboratories. The problem is a high pressure in the OPA solvent system in the PCX - 5200. How can we solve the problem, any sugestion?
Thanks in advance for your help!

Nelson Barros

There are a couple possibilities, but the most likely is the restrictor column is fouled. Take the OPA gauge panel off, and find the thing that looks like a small HPLC column, and take it out. You can try cleaning it by washing with water then acetone. Don't take it apart. If that fails, replace it. Also, check that the flow rate of OPA reagent is correct.

Call Pickering. Their tech support is pretty good. (I used to work there.)
Mark Tracy
Senior Chemist
Dionex Corp.

Thanks Mark for the help, i will try to clean that restrictor!
3 posts Page 1 of 1

Who is online

In total there are 12 users online :: 3 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Majestic-12 [Bot], Semrush [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry