alpha » Mon Sep 10, 2007 3:33 pm
Peter, by saying conditions have not changed, I only meant no new products have been introduced. For example, I have always used Hamilton liners and didn't switch from Restek. I do weekly, if not daily, maintenance on my instrument. Liners and septa get changed almost daily....column gets clipped ~ weekly, if not sooner. And gold seal is changed monthly, if not sooner.
AICMM, we have noticed that using wool can increase the tailing of some compounds. This becomes a problem because of Method 8270 witch has tailing criteria that has to be met. As for the flow rate, that is not an option at the moment. If we do that, it will put us weeks behind since new MDLs, calibration, etc will have to be done. that is a possiblity for the future when business slows down.
Thanks for the helpful advice.