Hi can someone please tell me how to calculate the number of column volumes of mobile phase passing through the column?
Thanks
Liv

Dear Olivia

The number of column volumes can be calculated knowing the total volume of the mobile phase that you use to purge the column and the column void volume (specified by manufacturer, see the flip of paper that comes with your column).

Number of column volumes = (purge volume of mobile phase)/(column void volume)

If you purge your column with 10 mL mobile phase (e.g. 10 minutes with 1mL/min flow) and your column void volume is 0.8mL then number of column volumes (that you purged you column after 10 min with 1mL/min flow) is 10mL/0.8mL = 12.5 column volumes.

Or, if you want to purge your column with 20 column volumes of mobile phase and your column void volume is 0.8mL, then total volume of mobile phase that should purge your column is 20x0.8mL = 16mL (purging 16 minutes with 1mL/min flow)

Best regards

thank you zokitano
3 posts Page 1 of 1

Who is online

In total there are 11 users online :: 2 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry