Celiagg » Tue Sep 11, 2007 7:41 pm
Hello Rubén
1.Frist at all, I would meassure the flow (better, with a glass flask of 10mL and a chromometer). Check that it is constant several times, and if you work in gradient way, with each of two phases separately, ( if each solvent goes by different valve...).
2.Once you have descarte the pump, I'll try to degass solvents properly with Helio bubles during 1/2 hour at least, to guarantee that are no bubbles of air the reason of migrating retention times.
3.After this, I would try to fix temperature. Some analyte's tr change a lot with temperatures changes of 3 ºC or less...
4. I would be sure of stability of the sample in the solvent mix that you use to dissolve it to inject . I'd make the same synthetic sample several times and after, I'd inject them with 1 hour between them: I mean, first fresh , second that is an hour from preparation...and so. Sorry, my Englihs is too old...
If your analytes are reacting between them, the product may be more absortive in that lambda. So areas may change.
5. At the end, I would try to investigate how long is since someone put the lamp into the detector. It may be low of energy, and make fluctuations in light. emission.
6. Are you using fast HPLC columns? If they are not well packed, it may make problems. like double peaks and losts of stationary phase. But it is not probable ...
7.
You must be sure that you mixture well th mobile phase before start .
And Thats all i can say...you can write to me at
cegrunder@yahoo.es
Si quieres comunicarte en español, no hay problema...
Regards