Greetings
Is anyone out there separating permanent gases with PLOT columns? What kind of lower detection limits can you get to?
I am specifically using Varian's dual Select Permanent Gases / CO2 (MolSiv5A PLOT + PorabondQ PLOT) but am not happy with O2 sensitivity. Can anyone recomend a better set of columns, or offer advice how to get down to low ppb levels here?
Many thanks-
:scratch:
PS I have eliminated the possibility of an O2 leak in my system. O2 seems to be physisorbing to my columns at these very low levels, losing sensitivity.

Your results are reasonable.

What do you mean when you say low ppb levels ?

1, 10, 100 ppb? How LOW do you need to GO ?

Depending upon detector used you may have to condition the analysis by injecting an amount of oxygen between analyses at some rate.

Micropacked columns with fused silica coated SS tubes may be necessary due to the larger amount of sample that can be injected, of course, conditioning spikes may still be necessary.

best wishes,

Rod
2 posts Page 1 of 1

Who is online

In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry