Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
9 posts
Page 1 of 1
*Due to a new wheat grain test, another lab analyst and me will be using the same HPLC; however, we use different types of columns (I, an RX Sil and her, a C-18) and different mobile phase formulations (I, 96%hexane:4%ethyl ether and her, 15% MeOH: 85% aqueous). So, we will be switching out columns and mobile phase fairly often during September. Our question is: when the column is not in use, what is the proper procedure for unhooking it and storing it until it is used again in three days time?
Jumpshooter
-
- Posts: 212
- Joined: Fri Jul 20, 2007 7:51 pm
(continuation of first post!)... and also, are any special procedures in place to re-start the use of either column type once the system is switched back over? Do we need to rinse it with water or some solvent prior to using both the LC and the respective columns?
Jumpshooter
-
- Posts: 267
- Joined: Tue Oct 12, 2004 9:35 pm
You can leave your column stored in the hexane/ether solvent, but the key point is to be sure that you flush your system with a solvent that is miscible with both hexane and methanol when switching over, such as chloroform. You will have to be more careful when switching back to your normal phase method to avoid getting water on your silica colum.
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
^ I agree, except, I'd do a pretty thorough 100% MeOH flush before repeating with the MP of the day (don't forget the injector...)
Thanks,
DR

DR

-
- Posts: 212
- Joined: Fri Jul 20, 2007 7:51 pm
How do you be sure that the Injector has been flushed thoroughly with MeOH? At this point, what I do is remove the analytical column and use a coupler to attach the inlet and outlet metal tubing together, then pump through MeOH or Chloroform at 0.5 ml per min for 30 mins. How do I also get the injector to be flushed by this washing step?
Jumpshooter
-
- Posts: 418
- Joined: Mon Aug 15, 2005 3:06 pm
Does your injector have a purge function? I've not seen an autosampler that didn't, so I'd expect yours would. If so, then running a standard purge should be OK.
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
One or two high volume blank injections at the start of 1st set after each changeover will confirm if things ahve been adequately flushed.
Thanks,
DR

DR

-
- Posts: 212
- Joined: Fri Jul 20, 2007 7:51 pm
OK, thankie DR---I typically do a 50 uL inject my Reagent Blank at position # 0, at the starte of each analytical run--but I will do two such injects after changing over from her wheat assay runs.
Jumpshooter
9 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google Feedfetcher and 16 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google Feedfetcher and 16 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science