Does anyone use Harvard PHD 22/2000 pump for post column derivatization usage? How about the pulse during the flow switching? Can it be eliminated by adding dampener?

Thank you.

Syringe pumps do not need pulse dampers. In fact, pulse dampers require at least 500 psi of back pressure before they do anything at all, and few syringe pumps have both the pressure rating and volume capacity to operate in that range. (That is why the Pickering syringe pump is so costly, it is special-made for that operating region.) For your application, you do not need a long reaction loop, nor high temperatures, so the operating pressure will be very low.

What you will want to do is to have some relay controls to refill the syringe at the beginning of each analysis. Then start the syringe pump, wait for a stable baseline, and start the chromatography.
Mark Tracy
Senior Chemist
Dionex Corp.
2 posts Page 1 of 1

Who is online

In total there are 7 users online :: 3 registered, 0 hidden and 4 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google [Bot], Semrush [Bot] and 4 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry