A. Ignore microlitre syringe – establish traceability by gas phase spiking if you can.
B. Don't calibrate - trust the syringe manufacturer markings or certificate.
C. Follow any ISO/ASTM/EPA etc. methods – if there are any.
D. Weigh the syringe before and after spiking the (cool) intermediate sorbent trap
E. Weigh liquid into a flask sealed with a septum.
F. Other
A few restrictions before opting C-F.
Mercury as the calibrant liquid is not allowed.
The syringe calibrant liquid must be the same as the dilution solvent for spiking the sorbent trap.
The dilution solvent cannot be semivolatile. It must be capable of being purged from the sorbent trap before analysis.
The uncertainty of the mean delivered volume must be not more than 2% (95% confidence.
The repeatability of the delivered volume must be not more than 1% (95% confidence).