JA » Fri Sep 14, 2007 11:01 pm
So we can state that for maximum retention in RPLC one uses a buffer with pH 2 or more units away (in the appropriate direction) from the analyte pKa to confer neutrality. These pKa values are, however, not readily available under the conditions found inside the HPLC column, i.e. in water-organic mixtures, and thus one may never know if ionisation is completely suppressed at a given mobile phase pH.
With the information given so far, the mobile phase pH suitable for your compound separation is answerable only by experimentation. Ionisation does not necessarily have to be suppressed to achieve a good result but I would agree that the more retention you've got, the more possibility of success there is.
Given sound working practice, there is no need to shy away from working at or close to the pKa of the compound. Good peak shape can be achieved along with a degree of retention which represents a balance of the ionised and unionised retention times.
A couple of threads for reference,
http://www.sepsci.com/chromforum/viewtopic.php?t=4751
http://www.sepsci.com/chromforum/viewtopic.php?t=5549