My A mobile phase is 5mM Ammonium Formate in Water.
The B mobile phase is Methanol with 5mM Ammonium Formate.
My analyte has two major metabolites with MW 732 and 746.
I found a step gradient that beautifully separates the target analytes from the rest of the junk in the wheat sample.
I'm using a 3um 2.1mm X 50mm dC-18 column.
The empirically derived step gradient is as follows.
Start off with
10% B at time 0
10% B to 10 min
80% B at 11 min
80% B to 20 min
95% B at 21 min
95% B to 25 min
10% B at 26 min
10% B to 30 min
end of run.
I full scan mode the some stuff washes off by 2.5 minutes,
other substantial unknowns elute 14-15.5 minutes, then big blob of eluents are observed 25-29min.
In MRM mode, my two analytes appear at 19.3 and 20.5 minutes.
I reinject a known sample 11 times in a row and had rock solid retention times. Peak areas varied from the initial injection within 5%. I used a known standard that was in 60% methanol/40% water.
Granted my LC method is a bit unconventional, it was just a test run to see what would come off in the aqueous and the organic phases and when.
I would use this method for our work, my attempts so far to ramp the gradient just hasn't produced the same results. I would therefore go ahead and use it as is. However, my colleague doesn't like step gradients and I should strictly use and gradient ramp. I understand the reasoning, but in this case the results are good.
Would you ever use a step gradient? Steps produce an effect that is different than ramps that in my case is desireable, IMHO.
Your comments are welcome.