Detector SN4000 problems

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi helpers
I have a problem with the detector. I am reading at 3 wavelength, 234, 280 and 300. and my chromatograms are not starting from 0 mAU, they start from 18, 210 and 200 mAU. I did the `review run for about 30 min. What could be the problem and how I could resolve it.

Thanks very much
Rachid Ganga

I assume you are autozeroing the detector sometime before the start of run. Check that the detector signal is actually stable at the time that the autozero is executed.
Mark Tracy
Senior Chemist
Dionex Corp.

I assume you are autozeroing the detector sometime before the start of run. Check that the detector signal is actually stable at the time that the autozero is executed.
...stable at the lowest wavelength...
Thanks,
DR
Image
3 posts Page 1 of 1

Who is online

In total there are 19 users online :: 2 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry