Recently I purchased a semi-prep flow cell for the DAD, the standard flow cell is 10 mm - 120 bar pressure.
I purchased a 0.3 mm - 20 bar pressure flow cell.
I have been given conflicting advice. Some people argue the 20 bar is the max pressure for the whole system, and that I would have to change the ID of my tubing throughout the system to decrease the back pressure.
Others suggest the HPLC system is an open system, and so if I determine the pressure without the flow cell in series, and then reconnect it, I can determine the pressure across/in the flow cell which should be less than 20 bar.
Is there a way of monitoring this? Should I install an in-line pressure gauge on the other side of the detector?
OR, can I tell by the increase of the pressure registering on the system (the sensor is usually at the pumps). The problem here is: If I am loading a large amount of sample, and have a gradient of three different solvents, then the pressure is going to vary according to the composition of the gradient, and, as the sample passes through the column.
If I have the pressure monitor detection on, and see a line representing the pressure this may help except I may have different samples, and mobile phases etc. and therefore do not know what it should look like.
What I am asking is: Once I have determined the pressure difference with the flow cell in line (i.e. the pressure difference across the flow cell) and that it is less than 20 bar , do I just assume that since the tubing throughout the system is:0.17 mm ID, but the tubing at the flow cell inlet is:0.5 mm, and at the outlet is:0.8 mm, that the pressure should be OK in the flow cell?
What if, there is a blockage at the fraction collector overnight, could this cause back pressure in the flow cell and cause it to crack?