I'm not really 'experimenting' at all. My company produces a software platform for HPLC Method Development and Method Validation experimentation using DoE methodology. After you have a design with your input variables, we can create a sequence/sample set and associated methods in several data systems. The idea is to be able to run multi-column, multi-solvent, multi-variable designs in a walk-away mode.
We also import results data and do automated analysis, reporting and optimization.
I'm just the guy who is (recently) in charge of handling data exchange with the various instrument control programs like Galaxie, Empower, TotalChrom, etc. In the coming weeks I will be implementing connections with EZ Chrom and Chromeleon as well.
You can find out more here:
www.smatrix.com
Chris
I am curious as to how you are experimenting - when you say you do not have direct access to the instruments. Are you using Empower to work remotely?
Uwe has a very important point - the temperature readout doesn't tell you the temperature inside the column - I would suggest about 10 mins but the only way to check is experimentally ie do repeat injections until the retention times are stable. If you are considering temperatures above 40C you need to think about preheating the mobile phase.