3 posts Page 1 of 1
Dear ALL

We are analysing the acetic acid as residual solvent in one of the API.
We are not getting good peak shape at std. concentration,while injecting high concentration peak shape is ok.
can any one suggest solution.

regards

AKJ
Amit Kuamr Jain

Dear Amit

Please give us more information about the chromatographic system, carrier, flow, column, type of API etc.

Regards

To measure acetic acid by GC and from a pharmaceutical product matrix is not trivial.

The fact that you are getting a peak at all at low levels is commendable.

Any active site on your column, detector, or in your injection port will cause absorption of the acetic acid. Replacement of your injection liner often and use of the proper phase for your column will improve your analysis.

Acetic acid is usually done by ion chromatography or by HPLC.

There are many published methods and companies that sell HPLC columns that will be happy to assist you.

best wishes,

Rod
3 posts Page 1 of 1

Who is online

In total there are 21 users online :: 2 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry